The Moving Coil Galvanometer

Moving Charges and Magnetism · CBSE

Theory

Principle

  • A current-carrying coil placed in a magnetic field experiences a torque, and the magnitude of that torque depends on the strength of the current. The torque turns the coil until a spring's restoring torque balances it, and the resulting deflection measures the current.
  • Equivalently: the torque tends to rotate the coil so that the magnetic flux through it is maximum.

Construction, described for the diagram

  • A rectangular coil of many turns of insulated wire wound on a light non-magnetic frame, suspended between the poles of a strong permanent magnet.
  • The pole pieces are cut concave (cylindrical), and a cylindrical soft-iron core sits inside the coil, so the field in the gap is *radial* — always in the plane of the coil.
  • A phosphor-bronze suspension strip above carries the current in and provides the restoring torque; a loose spring below carries the current out.
  • A pointer attached to the suspension moves over a graduated scale. For use as a detector the pointer's zero sits at the middle of the scale, so the deflection can go either way depending on the direction of the current.
  • The coil is wound on a metallic frame so that eddy currents induced in it damp the motion electromagnetically — the instrument is dead-beat and settles without oscillating.

Working — the derivation

  • Torque on the coil of NN turns, area AA, carrying current II in field BB: τ=NIABsin⁡θ\tau = NIAB\sin\theta, θ\theta being the angle between the normal to the coil and B⃗\vec{B}.
  • Because the field is radial, the plane of the coil is parallel to B⃗\vec{B} in every position, so θ=90∘\theta = 90^\circ and sin⁡θ=1\sin\theta = 1 throughout the rotation: τ=NIAB\tau = N I A B
  • The spring provides a counter torque kϕk\phi (with kk the torsional constant, the restoring torque per unit twist, in N m rad−1\text{N m rad}^{-1}; some CBSE answers write it as CC). At the steady deflection ϕ\phi, kϕ=NIAB  ⇒  ϕ=(NABk)Ik\phi = N I A B \;\Rightarrow\; \phi = \left(\dfrac{N A B}{k}\right) I
  • The bracket is a constant for a given galvanometer, so ϕ∝I\phi\propto I: the deflection is directly proportional to the current and the scale is linear (uniform) and can be calibrated evenly.

Why the radial field, and why the soft-iron core

  • Radial field: it keeps the plane of the coil parallel to B⃗\vec{B} at every angular position, so sin⁡θ=1\sin\theta = 1 always. This gives the maximum torque in every position and, crucially, makes ϕ\phi strictly proportional to II — a linear scale. Without it τ\tau would vary as sin⁡θ\sin\theta and the scale would be crowded and non-uniform.
  • How it is achieved: by cutting the magnetic pole pieces into concave (cylindrical) faces, and by placing a cylindrical soft-iron core inside the coil.
  • Soft-iron core: it has high permeability, so it concentrates the field lines, making the field both radial in the gap and considerably stronger — which increases the sensitivity.

Sensitivities

  • Current sensitivity: the deflection per unit current. IS=ϕI=NABkI_S = \dfrac{\phi}{I} = \dfrac{N A B}{k}
  • Increase it by increasing the number of turns NN, increasing the area AA, using a stronger radial field BB, or using a suspension of smaller torsional constant kk. Increasing NN is the manufacturer's convenient route.
  • Voltage sensitivity: the deflection per unit voltage across the galvanometer, with RR its coil resistance. VS=ϕV=NABkR=ISRV_S = \dfrac{\phi}{V} = \dfrac{N A B}{k R} = \dfrac{I_S}{R}
  • Why increasing current sensitivity need not increase voltage sensitivity: doubling NN doubles ISI_S, but it also roughly doubles RR, since twice the length of wire is used. In VS=IS/RV_S = I_S/R the two changes cancel and the voltage sensitivity is unchanged — or may even fall if RR grows faster than ISI_S.
  • Numerical version of the same point: if ISI_S rises by 20%20\% and RR by 25%25\%, then VS′=1.2IS1.25R=2425VSV_S' = \dfrac{1.2 I_S}{1.25 R} = \dfrac{24}{25}V_S, a decrease of 4%4\%.

Why a galvanometer is not itself an ammeter

  • It is extremely sensitive: full-scale deflection occurs for currents of the order of microamperes, so ordinary circuit currents would drive it far past the scale.
  • It must be connected in series to measure current, and its resistance is large enough to change the very current it is measuring.
  • It can be used as a *detector* (does current flow, and which way?), as in a Wheatstone bridge, and it can be converted into an ammeter or a voltmeter of a chosen range.

Conversion into an ammeter

  • Connect a small resistance rsr_s (the shunt) in *parallel* with the galvanometer, so most of the current bypasses the coil.
  • At full-scale current IgI_g through the galvanometer of resistance GG, and total current I0I_0 to be read, the shunt carries I0−IgI_0 - I_g and the two are at the same potential difference: IgG=(I0−Ig)S  ⇒  S=(IgI0−Ig)GI_g G = \left(I_0 - I_g\right) S \;\Rightarrow\; S = \left(\dfrac{I_g}{I_0 - I_g}\right) G
  • To multiply the range n=I0/Ign = I_0/I_g times, S=Gn−1S = \dfrac{G}{n-1}.
  • Resistance of the resulting ammeter: Reff=GSG+S=GIgI0R_{\text{eff}} = \dfrac{G S}{G + S} = \dfrac{G I_g}{I_0} — smaller than either GG or SS, and smaller still for a higher range. So the higher the range, the lower the resistance of the ammeter.
  • An ammeter must have a low resistance because it goes in series: a high resistance would raise the total circuit resistance and reduce the current being measured. An ideal ammeter has zero resistance.

Conversion into a voltmeter

  • Connect a large resistance RR in *series* with the galvanometer, so it draws only a tiny current, and connect the combination in parallel across the element.
  • For full-scale deflection at a potential difference VV: V=Ig(G+R)  ⇒  R=VIg−GV = I_g\left(G + R\right) \;\Rightarrow\; R = \dfrac{V}{I_g} - G
  • To multiply the range nn times, R=(n−1)GR = (n-1)G. The resistance of the finished voltmeter is G+R≃RG+R\simeq R, large.
  • A voltmeter must have a high resistance because it goes in parallel: it must draw negligible current so it does not disturb the potential difference it is reading (usually kept to under one per cent). An ideal voltmeter has infinite resistance.
  • Because the required modifications are quite different, the design that suits an ammeter does not suit a voltmeter.
FeatureAmmeterVoltmeter
Connectedin seriesin parallel
Extra resistancesmall shunt SS, in parallellarge RR, in series
ValueS=IgGI0−IgS = \dfrac{I_g G}{I_0 - I_g}R=VIg−GR = \dfrac{V}{I_g} - G
Range nn timesS=Gn−1S = \dfrac{G}{n-1}R=(n−1)GR = (n-1)G
Instrument resistanceGIgI0\dfrac{G I_g}{I_0}, lowG+RG+R, high
Ideal valuezeroinfinite

Worked numericals from the board papers

  • Ammeter of resistance 0.8 Ω0.8\ \Omega reading to 1.0 A1.0\ \text{A}, to read to 5.0 A5.0\ \text{A}: equal potential differences give 1×0.8=(5−1)R1\times0.8 = (5-1)R, so R=0.2 ΩR = 0.2\ \Omega.
  • Galvanometer G=16 ΩG = 16\ \Omega, Ig=4 mAI_g = 4\ \text{mA}, to read to 3 V3\ \text{V}: R=34×10−3−16=750−16=734 ΩR = \dfrac{3}{4\times10^{-3}} - 16 = 750-16 = 734\ \Omega in series.
  • Galvanometer G=15 ΩG = 15\ \Omega, Ig=4 mAI_g = 4\ \text{mA}, ammeter of range 00 to 6 A6\ \text{A}: S=4×10−36−0.004×15≈0.01 ΩS = \dfrac{4\times10^{-3}}{6-0.004}\times 15 \approx 0.01\ \Omega.
  • Galvanometer G=15 ΩG = 15\ \Omega, Ig=6 mAI_g = 6\ \text{mA}, range 00 to 6 A6\ \text{A}: S=6×10−3×155.994≈0.015 ΩS = \dfrac{6\times10^{-3}\times15}{5.994} \approx 0.015\ \Omega.
  • Series resistances 980 Ω980\ \Omega and 470 Ω470\ \Omega, the second halving the range: V1=Ig(G+980)V_1 = I_g(G+980), V1/2=Ig(G+470)V_1/2 = I_g(G+470), so 2(G+470)=G+9802(G+470) = G+980 and G=40 ΩG = 40\ \Omega.
  • Ranges 00–1 V1\ \text{V} with R1R_1 and 00–2 V2\ \text{V} with R2R_2: dividing the two equations gives 2(G+R1)=G+R22(G+R_1) = G+R_2, so G=R2−2R1G = R_2 - 2R_1.
  • Ranges VV with 2 kΩ2\ \text{k}\Omega and 2V2V with 5 kΩ5\ \text{k}\Omega: 2(G+2000)=G+50002(G+2000) = G+5000 gives G=1000 ΩG = 1000\ \Omega, then Ig=V/3000I_g = V/3000, and for range V/2V/2, V2=V3000(1000+R)\dfrac{V}{2} = \dfrac{V}{3000}(1000+R) gives R=500 ΩR = 500\ \Omega.

Results to carry away

  • kϕ=NIABk\phi = NIAB, ϕ=NABkI\phi = \dfrac{NAB}{k}I, ϕ∝I\phi\propto I.
  • IS=NABkI_S = \dfrac{NAB}{k}, VS=NABkR=ISRV_S = \dfrac{NAB}{kR} = \dfrac{I_S}{R}.
  • Ammeter: shunt in parallel, S=IgGI0−IgS = \dfrac{I_g G}{I_0-I_g}, effective resistance GIgI0\dfrac{G I_g}{I_0}.
  • Voltmeter: series resistance, R=VIg−GR = \dfrac{V}{I_g}-G.
  • Radial field for a linear scale; soft-iron core for a stronger radial field; metallic frame for electromagnetic damping.

Theory Questions Asked on This Subtopic

Real board questions. Answers are at the end of the page.

  1. 1

    Define the term 'current sensitivity' of a moving coil galvanometer.

    1 markCBSE Class 12 Physics — Very Short Answer · 2020, R · Section A, 1 markAnswer 1 ↓
  2. 2

    Write the underlying principle of a moving coil galvanometer.

    1 markCBSE Class 12 Physics — Very Short Answer · Delhi 2016 · Section A, 1 markAnswer 2 ↓
  3. 3

    An ammeter of resistance 0.8 Ω\Omega can measure a current upto 1.0 A. Find the value of shunt resistance required to convert this ammeter to measure a current upto 5.0 A.

    2 marksCBSE Class 12 Physics — SA · 2020 · Section B, 2 marksAnswer 3 ↓
  4. 4

    A galvanometer of resistance 16 Ω\Omega shows full scale deflection for a current of 4 mA. How will you convert it into a voltmeter to measure a voltage up to 3 V?

    3 marksCBSE Class 12 Physics — SA · 2020 · Section C, 3 marksAnswer 4 ↓
  5. 5

    (a) Briefly describe how the current sensitivity of a moving coil galvanometer can be increased. (b) A galvanometer shows full scale deflection for current IgI_g. A resistance R1R_1 is required to convert it into a voltmeter of range (0−1 V)(0-1\text{ V}) and a resistance R2R_2 to convert it into a voltmeter of range (0−2 V)(0-2\text{ V}). Find the resistance of the galvanometer.

    3 marksCBSE Class 12 Physics — SA · 2023 · Section C, 3 marksAnswer 5 ↓
  6. 6

    (a) Define current sensitivity of a galvanometer. Write its expression. (b) A galvanometer has resistance GG and shows full scale deflection for current IgI_g. (i) How can it be converted into an ammeter to measure current upto I0I_0 (I0>IgI_0 > I_g)? (ii) What is the effective resistance of this ammeter?

    3 marksCBSE Class 12 Physics — SA · 2020 · Section C, 3 marksAnswer 6 ↓
  7. 7

    (a) Briefly explain how a galvanometer is converted into an ammeter. (b) A galvanometer coil has a resistance of 15 Ω\Omega and it shows full scale deflection for a current of 4 mA. Convert it into an ammeter of range 0 to 6 A.

    2 marksCBSE Class 12 Physics — SA · AI 2019, U · Section B, 2 marksAnswer 7 ↓
  8. 8

    (a) Why do we use a shunt to convert a galvanometer into an ammeter? (b) A galvanometer of resistance 15 Ω\Omega shows a full scale deflection on the meter scale for a current of 6 mA. Calculate the value of the shunt resistance required to convert the galvanometer into an ammeter of range 0 - 6 A.

    2 marksCBSE Class 12 Physics — SA · AI 2019C, AI 2020C · Section B, 2 marksAnswer 8 ↓
  9. 9

    (a) Briefly explain how a galvanometer is converted into a voltmeter. (b) A voltmeter of a certain range is constructed by connecting a resistance of 980 Ω\Omega in series with a galvanometer. When the resistance of 470 Ω\Omega is connected in series, the range gets halved. Find the resistance of the galvanometer.

    3 marksCBSE Class 12 Physics — SA · AI 2019 · Section C, 3 marksAnswer 9 ↓
  10. 10

    Describe the working principle of a moving coil galvanometer. Why is it necessary to use (i) a radial magnetic field and (ii) a cylindrical soft iron core in a galvanometer? Write the expression for current sensitivity of the galvanometer. Can a galvanometer as such be used for measuring the current? Explain.

    5 marksCBSE Class 12 Physics — Long Answer · Delhi 2017, Cr · Section E, 5 marksAnswer 10 ↓
  11. 11

    State the principle of working of a galvanometer. A galvanometer of resistance GG is converted into a voltmeter to measure upto VV volts by connecting a resistance R1R_1 in series with the coil. If a resistance R2R_2 is connected in series with it, then it can measure upto V/2V/2 volts. Find the resistance, in terms of R1R_1 and R2R_2, required to be connected to convert it into a voltmeter that can read upto 2V. Also find the resistance GG of the galvanometer in terms of R1R_1 and R2R_2.

    3 marksCBSE Class 12 Physics — SA · 2016 · Section C, 3 marksAnswer 11 ↓
  12. 12

    (a) Why is the magnetic field radial in a moving coil galvanometer? Explain how it is achieved. (b) A galvanometer of resistance 'G' can be converted into a voltmeter that can read upto 2 V. Also find the resistance GG of the galvanometer in terms of R1R_1 and R2R_2.

    5 marksCBSE Class 12 Physics — Long Answer · Delhi 2015 · Section E, 5 marksAnswer 12 ↓
  13. 13

    (a) With the help of a diagram, explain the working of a moving coil galvanometer. Justify the necessity of using radial magnetic field in it. (b) A galvanometer can be converted into a voltmeter to measure upto (i) VV volt by connecting a resistance of 2 kΩ2\text{ k}\Omega in series with the galvanometer. (ii) 2V2V volt by connecting a resistance of 5 kΩ5\text{ k}\Omega in series with the galvanometer. Calculate the resistance that should be connected in series with the galvanometer to convert it into a voltmeter to measure upto V/2V/2 volt.

    5 marksCBSE Class 12 Physics — Long Answer · AI 2020C · Section E, 5 marksAnswer 13 ↓
  14. 14

    (i) With the help of a neat and labelled diagram, explain the principle and working of a moving coil galvanometer. (ii) What is the function of uniform radial field and how is it produced? (iii) Define current sensitivity of a galvanometer. How is current sensitivity increased?

    5 marksCBSE Class 12 Physics — Long Answer · Foreign 2016, Ev · Section E, 5 marksAnswer 14 ↓
  15. 15

    (a) Explain using a labelled diagram, the principle and working of a moving coil galvanometer. What is the function of (i) uniform radial magnetic field, (ii) soft iron core? (b) Define the terms (i) current sensitivity and (ii) voltage sensitivity of a galvanometer. Why does increasing the current sensitivity not necessarily increase voltage sensitivity? OR (a) Draw a labelled diagram of a moving coil galvanometer. Describe briefly its principle and working. (b) Answer the following: (i) Why is it necessary to introduce a cylindrical soft iron core inside the coil of a galvanometer? (ii) Increasing the current sensitivity of a galvanometer may not necessarily increase its voltage sensitivity. Explain, giving reason.

    5 marksCBSE Class 12 Physics — Long Answer · AI 2015 · Section E, 5 marksAnswer 15 ↓

Numericals

Every numerical set on this subtopic. Full working is at the end of the page.

  1. 16

    If an ammeter is to be used in place of a voltmeter, then we must connect with the ammeter a

    • (a)

      low resistance in parallel

    • (b)

      low resistance in series

    • (c)

      high resistance in parallel

    • (d)

      high resistance in series.

    1 markCBSE Class 12 Physics — Multiple Choice · Term I 2021-22 · Section A, 1 markAnswer 16 ↓
  2. 17

    Question No. 65 is Assertion (A) and Reason (R) type question. Given below are the two statements labelled as Assertion (A) and Reason (R). Select the most appropriate answer from the options given below. Assertion (A): Higher the range, lower is the resistance of an ammeter. Reason (R): To increase the range of an ammeter additional shunt is added in series to it.

    • (a)

      Both Assertion (A) and Reason (R) are true and Reason (R) is the correct explanation of the Assertion (A).

    • (b)

      Both Assertion (A) and Reason (R) are true, but Reason (R) is not the correct explanation of the Assertion (A).

    • (c)

      Assertion (A) is true, but Reason (R) is false.

    • (d)

      Assertion (A) is false and Reason (R) is also false.

    1 markCBSE Class 12 Physics — Multiple Choice · Term I 2021-22, 2021C, Ap · Section A, 1 markAnswer 17 ↓

Answers & Solutions

Work the questions above first, then check yourself here. Each answer is set out the way it should be written in the paper.

  1. Current sensitivity is defined as the deflection of the coil per unit current flowing in it, i.e. Is=θI=NABkI_s = \dfrac{\theta}{I} = \dfrac{NAB}{k} where NN is the number of turns, AA the area, BB the magnetic field strength, and kk the torsional constant of the suspension.

  2. When a current carrying coil is suspended in a uniform magnetic field, a torque acts on it, the magnitude of which depends on the strength of current. This torque tends to rotate the coil about the axis of suspension, so that the magnetic flux passing through the coil is maximum.

  3. Let the shunt resistance be RR. So, Vshunt=VammeterV_{shunt} = V_{ammeter} (parallel combination), giving: 1×0.8=(5−1)×R1\times0.8 = (5-1)\times R 0.8=4R⇒R=0.2 Ω0.8 = 4R \Rightarrow R = 0.2\ \Omega

  4. To convert it into a voltmeter, a large resistance RR is connected in series with the galvanometer. ∴R=VIg−RG=34×10−3−16\therefore R = \dfrac{V}{I_g} - R_G = \dfrac{3}{4\times10^{-3}} - 16 R=750−16=734 ΩR = 750 - 16 = 734\ \Omega

  5. (a) The current sensitivity is given by Is=θI=NABkI_s = \dfrac{\theta}{I} = \dfrac{NAB}{k}. Current sensitivity can be increased when: (i) number of turns NN is increased; (ii) area of coil AA is increased; (iii) magnetic field BB should be radial (and made as strong as possible, e.g. using a stronger permanent magnet). (b) 1 V=Ig(G+R1)1\text{ V} = I_g(G+R_1) ...(i) 2 V=Ig(G+R2)2\text{ V} = I_g(G+R_2) ...(ii) Dividing (ii) by (i): 21=G+R2G+R1\dfrac{2}{1} = \dfrac{G+R_2}{G+R_1} ⇒2G+2R1=G+R2⇒G=R2−2R1\Rightarrow 2G + 2R_1 = G + R_2 \Rightarrow G = R_2 - 2R_1

  6. (a) Current sensitivity is defined as the deflection of coil per unit current flowing in it, i.e. Is=θI=NABkI_s = \dfrac{\theta}{I} = \dfrac{NAB}{k}. (b)(i) A galvanometer can be converted into an ammeter of given range by connecting a suitable low resistance SS called shunt in parallel to the given galvanometer, whose value is given by S=(IgI0−Ig)GS = \left(\dfrac{I_g}{I_0-I_g}\right)G where IgI_g is the current for full scale deflection of galvanometer, I0I_0 is the current to be measured by the galvanometer and GG is the resistance of galvanometer. (ii) The effective resistance of this ammeter (galvanometer GG parallel with shunt SS) is Reff=GSG+SR_{eff} = \dfrac{GS}{G+S}, which works out to Reff=GIgI0R_{eff} = \dfrac{G I_g}{I_0} (using S=IgGI0−IgS = \dfrac{I_g G}{I_0-I_g}, the parallel combination simplifies to this low value, always less than SS and GG individually).

  7. (a) A galvanometer can be converted into an ammeter of given range by connecting a suitable low resistance SS called shunt in parallel to the given galvanometer, whose value is given by S=(IgI0−Ig)GS = \left(\dfrac{I_g}{I_0-I_g}\right)G, where IgI_g is the current for full scale deflection of galvanometer, I0I_0 is the current to be measured by the galvanometer and GG is the resistance of galvanometer. In order to increase the range of an ammeter nn times, the value of shunt resistance to be connected in parallel is S=G/(n−1)S = G/(n-1). (b) For I0=6I_0 = 6 A: S=(IgI0−Ig)G=4×10−36−0.004×15S = \left(\dfrac{I_g}{I_0-I_g}\right)G = \dfrac{4\times10^{-3}}{6-0.004}\times15 S≈4×10−36×15=0.01 ΩS \approx \dfrac{4\times10^{-3}}{6}\times15 = 0.01\ \Omega

  8. (a) As the ammeter is connected in series in the circuit, therefore, its resistance must be very low so that the circuit current is not affected. A very low shunt resistance makes the effective resistance of galvanometer (and hence the ammeter) very low. (b) S=IgGI−Ig=6×10−3×156−6×10−3S = \dfrac{I_g G}{I-I_g} = \dfrac{6\times10^{-3}\times15}{6-6\times10^{-3}} S=6×10−3×155.994≈0.015 ΩS = \dfrac{6\times10^{-3}\times15}{5.994} \approx 0.015\ \Omega

  9. (a) A galvanometer can be converted into a voltmeter of given range by connecting a suitable resistance RR in series with the galvanometer, whose value is given by R=VIg−GR = \dfrac{V}{I_g} - G, where VV is the voltage to be measured, IgI_g is the current for full scale deflection of galvanometer and GG is the resistance of galvanometer. Voltmeter is a high resistance instrument and it is always connected in parallel with the circuit element across which potential difference is to be measured. An ideal voltmeter has infinite resistance. In order to increase the range of voltmeter nn times, the value of resistance to be connected in series with galvanometer is R=(n−1)GR = (n-1)G. (b) V1=Ig(R+980)V_1 = I_g(R+980) and V2=Ig(R+470)V_2 = I_g(R+470) Also, V2=V1/2V_2 = V_1/2 (given the range gets halved with the smaller series resistance) ⇒2(R+470)=R+980⇒2R+940=R+980⇒R=40 Ω\Rightarrow 2(R+470) = R+980 \Rightarrow 2R + 940 = R + 980 \Rightarrow R = 40\ \Omega

  10. Working principle: It works on the principle that a current-carrying coil placed in a magnetic field experiences a torque, the magnitude of which depends on the strength of current. (i) Radial magnetic field: To maintain the plane of the coil always parallel to the field B⃗\vec{B} and to have maximum torque at every orientation of the coil (so that θ=90°\theta = 90° always and sin⁡θ=1\sin\theta = 1 throughout the rotation), giving a uniform, linear response. (ii) A cylindrical soft iron core: This has high permeability and it intensifies the magnetic field and hence increases the sensitivity of the galvanometer. Current sensitivity, IS=θI=NBAkI_S = \dfrac{\theta}{I} = \dfrac{NBA}{k} No, a galvanometer as such cannot be used directly for measuring current in amperes, because it is an extremely sensitive instrument (full-scale deflection occurs even for a very small current) and it does not have a calibrated current scale of its own; it can, however, be used to detect current or, with a shunt, converted into an ammeter of a specific range. By measuring the deflection produced in the galvanometer coil (relating θ\theta to II via its known sensitivity), one can in principle obtain the current, but not as a ready-made ammeter reading.

  11. Principle: A current carrying coil placed in a magnetic field experiences a torque, the magnitude of which depends on the strength of current. Galvanometer as a voltmeter — Case (I): R1=VIg−GR_1 = \dfrac{V}{I_g} - G ...(i) Case (II): R2=V2Ig−GR_2 = \dfrac{V}{2I_g} - G ...(ii) Case (III), for range 2V2V: R3=2VIg−GR_3 = \dfrac{2V}{I_g} - G ...(iii) IgI_g = current through galvanometer which is fixed. From eqns. (i) and (ii): R1−R2=V2IgR_1 - R_2 = \dfrac{V}{2I_g}, so G=R1−2R2G = R_1 - 2R_2 Putting these values in eqn. (iii), we get R3=4(R1−R2)−(R1−2R2)=3R1−2R2R_3 = 4(R_1-R_2) - (R_1-2R_2) = 3R_1 - 2R_2

  12. (a) Magnetic field is radial in moving coil galvanometer so that the plane of the coil always lies in the direction of the magnetic field (i.e. θ=90°\theta = 90° is maintained at every orientation), giving maximum and constant torque, so that the scale is linear. A radial magnetic field is produced by (i) properly cutting the magnetic pole pieces in the shape of concave faces, and (ii) using a soft iron core within the coil. (b) V=(R+G)IgV = (R+G)I_g ...(i) V′=(R′+G)IgV' = (R'+G)I_g ...(ii) From equations (i) and (ii): V′V=R′+GR+G=12⇒R′=R2−G2\dfrac{V'}{V} = \dfrac{R'+G}{R+G} = \dfrac{1}{2} \Rightarrow R' = \dfrac{R}{2} - \dfrac{G}{2} (This gives the resistance needed to change the range from 00 to VV down to 00 to V/2V/2, in terms of the original series resistance RR and galvanometer resistance GG.)

  13. (a) Working: when a current flows through the coil, a torque acts on it, τ=NIAB\tau = NIAB (taking sin⁡θ=1\sin\theta=1 since the field is radial). This magnetic torque tends to rotate the coil, and a spring provides a counter (restoring) torque kϕk\phi that balances the magnetic torque, resulting in a steady angular deflection ϕ\phi. In equilibrium, kϕ=NIAB⇒ϕ=(NABk)Ik\phi = NIAB \Rightarrow \phi = \left(\dfrac{NAB}{k}\right)I, i.e. ϕ∝I\phi \propto I. Since deflection is directly proportional to current, we get a linear scale. The uniform radial magnetic field keeps the plane of the coil always parallel to the direction of the magnetic field for every orientation, ensuring sin⁡θ=1\sin\theta = 1 throughout the coil's rotation, which is what makes the scale linear (justifying its necessity). (b) Case (i): V=Ig(G+2kΩ)V = I_g(G+2\text{k}\Omega) ...(i) Case (ii): 2V=Ig(G+5kΩ)2V = I_g(G+5\text{k}\Omega) ...(ii) Dividing (ii) by (i): 2=G+5000G+2000⇒2G+4000=G+5000⇒G=1000 Ω2 = \dfrac{G+5000}{G+2000} \Rightarrow 2G+4000 = G+5000 \Rightarrow G = 1000\ \Omega From (i): Ig=VG+2000=V3000I_g = \dfrac{V}{G+2000} = \dfrac{V}{3000} For range V/2V/2: let series resistance be RR. V2=Ig(G+R)=V3000(1000+R)\dfrac{V}{2} = I_g(G+R) = \dfrac{V}{3000}(1000+R) ⇒1500=1000+R⇒R=500 Ω\Rightarrow 1500 = 1000+R \Rightarrow R = 500\ \Omega

  14. (i) Principle and working: A galvanometer is used to detect current in a circuit. When current II is passed in the coil, torque τ\tau acts on the coil, given by τ=NIABsin⁡θ\tau = NIAB\sin\theta, where θ\theta is the angle between the normal to plane of coil and the magnetic field of strength BB, and NN is the number of turns in the coil. When the magnetic field is radial (as in the case of cylindrical pole pieces and soft iron core), then in every position of coil, the plane of the coil is parallel to the magnetic field lines, so that θ=90°\theta = 90° and sin⁡90°=1\sin90° = 1. Deflecting torque, τ=NIAB\tau = NIAB. If CC is the torsional rigidity of the wire and ϕ\phi is the twist of suspension strip, then restoring torque =Cϕ= C\phi. For equilibrium, deflecting torque = restoring torque, i.e. NIAB=Cϕ⇒ϕ=(NABC)INIAB = C\phi \Rightarrow \phi = \left(\dfrac{NAB}{C}\right)I, i.e. ϕ∝I\phi \propto I. Deflection of coil is directly proportional to current flowing in the coil and hence we can construct a linear scale. (ii) The uniform radial magnetic field keeps the plane of the coil always parallel to the direction of the magnetic field, i.e. the angle between the plane of the coil and the magnetic field is zero (equivalently, θ=90°\theta=90° between normal and field) for all orientations of the coil. It is produced by using cylindrical pole pieces (concave in shape) together with a soft iron cylindrical core placed within the coil. (iii) Current sensitivity of a moving coil galvanometer is defined as the deflection of coil per unit current passed through it: IS=θI=NBAkI_S = \dfrac{\theta}{I} = \dfrac{NBA}{k}, where NN is the number of turns, AA is the area of the coil, BB is the magnetic field strength of the poles and kk is the spring (torsional) constant of the suspension wire. A convenient way to increase current sensitivity is to increase the number of turns NN (or the area AA, or use a stronger magnet for larger BB, or a suspension of lower torsional constant kk).

  15. (a) Principle and working: When current (II) is passed in the coil, torque τ\tau acts on the coil, given by τ=NIABsin⁡θ\tau = NIAB\sin\theta, where θ\theta is the angle between the normal to plane of coil and the magnetic field of strength BB, NN is the number of turns in a coil. When the magnetic field is radial, as in the case of cylindrical pole pieces and soft iron core, then in every position of coil, the plane of the coil is parallel to the magnetic field lines, so that θ=90°\theta = 90° and sin⁡90°=1\sin90° = 1. Deflecting torque, τ=NIAB\tau = NIAB. If CC is the torsional rigidity of the wire and θ\theta is the twist of suspension strip, then restoring torque =Cθ= C\theta. For equilibrium, deflecting torque = restoring torque, i.e. NIAB=CθNIAB = C\theta. ∴θ=NABCI\therefore \theta = \dfrac{NAB}{C}I, i.e. θ∝I\theta \propto I. Deflection of coil is directly proportional to current flowing in the coil, hence we can construct a linear scale. (i) The uniform radial magnetic field keeps the plane of the coil always parallel to the direction of the magnetic field, i.e. the angle between the plane of the coil and the magnetic field is zero for all orientations of the coil. (ii) The cylindrical soft iron core, when placed inside the coil of a galvanometer, makes the magnetic field stronger and radial in the space between it and the pole pieces, such that whatever be the position of the coil, the magnetic field is always parallel to its plane — this also increases sensitivity. (b) Current sensitivity: IS=NBAkI_S = \dfrac{NBA}{k}. Voltage sensitivity, VS=NBAkRV_S = \dfrac{NBA}{kR}, where RR is the resistance of the wire (coil). Voltage sensitivity = Current sensitivity / RR. Thus, on increasing the current sensitivity (e.g. by increasing NN), voltage sensitivity may or may not increase, because increasing NN also increases the resistance RR of the coil by a similar (or greater) proportion, and the two changes can offset each other in the ratio IS/RI_S/R.

  16. Answer: (d) high resistance in series. As the ammeter is connected in place of a voltmeter, its resistance should be high (a voltmeter is ideally a high-resistance instrument). So, we have to connect a high resistance in series.

  17. Answer: (c) (A) is true, but (R) is false. When the resistance is lower, the range of ammeter is higher. So, Assertion (A) is true. To increase the range of ammeter, additional shunt is added in parallel (not in series) with the galvanometer. So, Reason (R) is false.